Print
Author Topic: NEC P221W w/ spectraviewII calibrated with 2 different i1display 2 pucks  (Read 4068 times)
a_krause
Jr. Member
**
Offline Offline

Posts: 52


« on: July 24, 2009, 12:13:30 AM »
ReplyReply

I just recently bought an NEC P221 monitor the other week and am overall pretty happy with it. I just purchased the spectraviewII software online and ran some tests. I share an office with another photographer and we both have gretag macbeth eye 1 display 2 pucks. I was curious to see how each puck calibrated the monitor using the spectraview software. With that said, the pucks give a very different contrast ratio and delta E value, from looking at the attached summaries, does it look like one of the pucks is broken or more accurate?

Logged

pherold
Full Member
***
Offline Offline

Posts: 133


« Reply #1 on: July 24, 2009, 12:45:11 PM »
ReplyReply

It's not uncommon for a profiling result to end with luminance values that are different than your aim.  Your puck # 1 at 134 came in under your aim of 140 and puck #2 was 141.  These differences in luminance might affect a lot of things.  The brightness of a display affects color perception more than many people think.  I don't believe that by itself is responsible for your higher delta E values for puck #1 though.

Have you run the pucks through the i1Diagnostic program yet?  X-Rite has a free download of a diagnostic program that you can plug an i1 device into, and it will run it through its paces & give you some numbers & tell you if it passes.  This is not always a guarantee that things are okay, but it's helpful.
Logged

-Patrick Herold
  Tech Support
www.chromix.com
Czornyj
Sr. Member
****
Offline Offline

Posts: 1445



WWW
« Reply #2 on: July 24, 2009, 01:19:49 PM »
ReplyReply

Quote from: a_krause
I just recently bought an NEC P221 monitor the other week and am overall pretty happy with it. I just purchased the spectraviewII software online and ran some tests. I share an office with another photographer and we both have gretag macbeth eye 1 display 2 pucks. I was curious to see how each puck calibrated the monitor using the spectraview software. With that said, the pucks give a very different contrast ratio and delta E value, from looking at the attached summaries, does it look like one of the pucks is broken or more accurate?

To my eye they both look odd - all P221W I had calibrated (about 6-7?) had R,B colorants pretty close to AdobeRGB R,B colorants.
I suppose it's a general problem with an i1 colorimetrer, that doesn't work well with wide-gamut panels.

Check "Factory measurements" in Edit>Preferences>ICC Profile>Source of primary color chromaticities for ICC Profile. It would also be perfect to calibrate the panel with i1pro, ColorMunki spectrophotometer, or custom NEC colorimeter.

« Last Edit: July 24, 2009, 01:31:02 PM by Czornyj » Logged